FYI - We are using the properly filtered mobile phases and high quality eluent components. The system is new and we have been following the manufacturers recommendations for proper column use.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
About the increasing of Head pressure:
I'm working on wines, liquors, honey and soft drink sugars content.
Flow: 1 ml/min; eluent 80:20 ACN/ Water plus 0.2% Triethylamine. Samples at 25°C Column 35°C RI detector 35°C
I've destroyed two column because the increasing pressure.
Now, at the end of the day I clean up the column as follows:
First change to 80:20 ACN/water without Tryethylamine,
Then change to 50:50 60 min
then change to 50:50 ACN/Water plus 0.1% Trifluoroacetic acid and hold 60 min
New change to 50:50 ACN/Water 30 min abd finally store the column in 80:20 ACN/Water
I'm working like this three months and pressure increase has been no significant (Initially 1175psi, nowadays 1250 psi)
The column is 250 x 4.6 mm
Perhaps it can help
Guillermo
Dear JamieAlison,
I am having the same proble but on an HPLC column, could you tell me how did you solve the problem, please?
Thanks in advance,
Carol
Thanks so much for the guidance, Wiley One. Very helpful!!nice to hear.
would you like to tell us how you solved your problem or what was the cause?