GC-MS Baseline drift downward

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Good morning, all.

I've got a Shimadzu QP-2010S unit with a column splitter to go to my MS and my FID. Recently, I did an overhaul on the whole system and replaced all column lines. When running samples and blanks, I'm seeing the MS baseline drift downward during a run. I do MS sweeps throughout the run, since I don't know what material I'm looking for and I don't want to miss anything by single ion monitoring. I'm currently running a blank of headspace air to see if it repeats, but I thought it might be a good idea to touch base with you all to see if you had any thoughts.

Thanks!
Rule number 2 of trouble shooting; a problem that only occurs once is not a problem.

Peter
Peter Apps
Thanks, Peter. Unfortunately, this is happening continuously. Thoughts on a cause for a downward drift to baseline?
How does your 207mz EIC look for column bleed.

How do you split your sample, eg two hole inlet ferrule, MicroFludic Splitter , Glass Y joint ,etc...
as more carrier gas may be getting in, diluting the ionization efficiency.

Have the relative responses changed on your test mixes or remained constant
from the early elutors to the last elutor, for both the MS and FID.

Alex
Thanks, Peter. Unfortunately, this is happening continuously. Thoughts on a cause for a downward drift to baseline?
Very likely to be connected with changes in gas flows through the two detector lines as the oven temperature increase changes gas viscosity. The details of how to fix it will depend on the plumbing.

Peter
Peter Apps
Thanks, Peter. Unfortunately, this is happening continuously. Thoughts on a cause for a downward drift to baseline?
Very likely to be connected with changes in gas flows through the two detector lines as the oven temperature increase changes gas viscosity. The details of how to fix it will depend on the plumbing.

Peter
Turns out, it was just that I didn't give the mass spec enough time to stabilize. After a weekend, the baseline was flat and stable, with no issues observed during a run. I shouldn't have been surprised, since the machine was off for a month or so.
6 posts Page 1 of 1

Who is online

In total there are 70 users online :: 0 registered, 0 hidden and 70 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 70 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry