Why would my pressure drop during bfb injection?
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
7 posts
Page 1 of 1
I did injector maintainance on a HP series 6890. Replaced septa, liner, and gold seal then my bfb abundance dropped from 1,200,000 to 120,000. I have been working for about a week now trying to figure what might have caused this but to no avail. I have changed the column, cleaned the source, replaced the filaments, checked the electron multiplier. The only thing I notice is that after shooting the bfb, as soon as I press start, the front inlet pressure drops from 26.0 to 22.0 then back up to 27.0 before stabilizing. Could this be the problem? If so, then what could be a possible solution? (oh i've also checked the flows and they all seem normal)
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
As soon as you have an exact 10 x difference in results the first thing to eliminate is a change in the data processing settings. Nevertheless your numbers are suspiciously rounded, so I suspect that you are providing us with approximations. Some real numbers would be helpful.
A pressure fluctuation just after injection can indeed be the cause of the problem. I suspect that it is going from standby to run as you press the start - there should be an option to go from standby into pre-run or whatever they call it.
Peter
A pressure fluctuation just after injection can indeed be the cause of the problem. I suspect that it is going from standby to run as you press the start - there should be an option to go from standby into pre-run or whatever they call it.
Peter
Peter Apps
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
CarltonPHX,
BFB by liquid injection or via purge and trap? Also, splitless or split?
Best regards,
AICMM
BFB by liquid injection or via purge and trap? Also, splitless or split?
Best regards,
AICMM
-
- Posts: 4
- Joined: Thu Apr 03, 2014 1:39 am
Peter,
Thanks for the reply, I will check the data processing settings and let you know if anything has changed. Also, you are correct, the numbers that I provided for the pressure fluctuations are rounded because they are not consistent.
AICMM,
Thank you for your reply. It is a liquid injection/splitless.
Thanks for the reply, I will check the data processing settings and let you know if anything has changed. Also, you are correct, the numbers that I provided for the pressure fluctuations are rounded because they are not consistent.
AICMM,
Thank you for your reply. It is a liquid injection/splitless.
-
- Posts: 4
- Joined: Thu Apr 03, 2014 1:39 am
I checked the settings and nothing has changed.
-
- Posts: 4
- Joined: Thu Apr 03, 2014 1:39 am
I replaced the chem trap, and replaced the liner and still no change..
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
When you checked the flows which flows did you check (split, septum purge, column ?) and how did you measure them ?
What happens if you press start and then inject ?
Peter
What happens if you press start and then inject ?
Peter
Peter Apps
7 posts
Page 1 of 1
Who is online
In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 29 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 29 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science