I am analyzing the vegetable oil for glycerides estimation on HPLC - C18 column.
Solvent system: A-methanol. B (50% hexane + 50%Isopropanol).
Gradient of 0 min - 90%A + 10% B to 15min - 50%A + 50% B.
Detector is UV/Vis at 212nm (the only detector available in my Lab.)
At 5-6 min of run the baseline increase very much to no peak can be seen after this area.
Please guide/help
No Water in the Mobile Phase?????
What you see is probably the UV cut off from your B mobile phase. Maybe a higher wavelength than 210nm is recommended.
Gerhard Kratz, Kratz_Gerhard@web.de
No Water in the Mobile Phase?????
What you see is probably the UV cut off from your B mobile phase. Maybe a higher wavelength than 210nm is recommended.
I doubt about the water contamination in isopropanol. The recommended wavelength is 205nm but below 210nm the UV/Vis detector gives "A/Z out of range error" therefore I am using 212 nm.
The nominal UV cutoffs for MeOH and iPrOH are both 205 nm. However:
- UV cutoff is the wavelength at which the absorbance goes to 1.0 AU, which is high for HPLC.
- The slopes of the UV spectra for those alcohols is fairly shallow in thst region, so that you can still have significant absorbance in the 210-215nm range.
- The quality of the solvents can vary significantly.
- Essentially, at 212 nm, you are on the ragged edge of usability.

If this were my problem, I would run UV spectra of both the A and B solvents to see what's going on. If the absorbances look OK, it may be time for a new detector lamp and/or flow cell cleaning and alignment.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry