- Posts: 45
- Joined: Sat Oct 12, 2013 9:58 am
I have trouble detecting the last 3 PAHs:
- Indeno[1,2,3-cd]pyrene
- Dibenzo[a,h]anthracene
- Benzo[ghi]perylene
My settings are:
GC-MS: ThermoQuest GCQ Plus - ion trap quadrupole
Column name: Phenomenex Zebron ZB-5MS
Composition: 5 % Phenyl 95 % Dimethylpolysiloxane
Size: 30m x 0.25mm x 0.25µm
Ramp: 40 °C - 4 min, 240 °C - 25 °/min, 310 °C - 5 °/min, 310 °C - 10 min
Injector temp: 300 °C
Columns flow: He, 40 cm/sec
Injection type: Splitless
Needle rest in injector: 3s
Injection volume: 1µl (tried with 2 µl and 3 µl)
Transfer line temp: 300 °C
Ionization chamber temp: 240 °C
Standard: MRC with 1 mg/l (1000 ppb)
Tried both full scan and SIM, for SIM i used 4 time windows, with the ions for every PAH (Parent ion, Quantization ion, Qualifying ion)
Starting with Naphtalene and ending with Crysene everything is beautiful. Big peaks, good shape, good separation.
Benzofluoranthene elutes together with Benzo[k]fluoranthene in a very ugly shape, broadened. Same ugly, broadened and noisy form for Benzo[a]pyrene. But still identifiable and quantifiable.
However Indeno[1,2,3-cd]pyrene, Dibenzo[a,h]anthracene and Benzo[ghi]perylene are nowhere to be found. I tried A LOT of settings, with multiplicator, AGC, waveform filtering, scans per second. No luck.
I shortened the column, changed septa, liner, with new ones... nothing. Every other peak is very beautiful and strong S/N wise. Not the last PAHs...
Calibration is flawless, good signal, multiplier is auto set on 1300. Pressure in system is at 40mTorr, no leaks, no contamination.
Is it possible to be a column fault ? If not, what can it be ? Any advice would be appreciated.
Best regards,
Vlad Popovici