HPLC Method for Quantification of Urea

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi all,

I wanted to know if any has used a specific method for quantifying urea.

Currently I am doing:

Dissolving pure urea in 90:10 Acetonitrile:Buffer (ammonium acetate, acetic acid, pH 5)

Method:

Mobile phase- 93% acetonitrile 7% buffer
Column- NH2
UV- 200, 210, 230
RI Detector
Inj Volume- 5 uL
Flow- 1 mL/min
Column Temp- 40

So far I have not even been seeing the urea. I have made up many concentrations and nothing seems to have an effect. There is a peak near 5 on the RID and DAD, but it never changes with different concentrations. The detector is not saturated and neither is the solution.

Any ideas or experience would be greatly appreciated.

Thanks!
"Determination of urea, allantoin and lysine pyroglutamate in cosmetic samples by hydrophilic interaction chromatography" - Ph. Dallet et al., J. Chromatogr. B, 742 (2000) 447.
The authors used a PolyHYDROXYETHYL A column (200x4.6-mm; 5-µm, 60-Å) with 6 mM triethylammonium phosphate (TEAP), pH 2.8, and varying amounts of ACN. It looks like they settled on 88% ACN. Detection was via absorbance at 200 nm.

Commenting on your method: Acetate ion has a much higher molar extinction coefficient than does urea at all of the wavelengths that you are monitoring. TEAP was used in the above paper because it's transparent at 200 nm. You can have a mobile phase that's transparent at low wavelengths or one that's volatile, but not both. An exception is one with TFA, but that antagonizes retention in the HILIC mode. When you're performing HILIC of urea, you need all the retention that you can get.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Thank you! I will look into that. I have seen that the NH2 column could cause quite a bit of baseline noise (which is present).

I may try to use a phosphate buffer instead of the ammonium acetate if I can't seem to pinpoint any other problems.
An Aminex HPX-87C column with RI will work for urea . Mobile phase : Water @80C 0.6 ml/min.

- Karen
I'm taking this from memory, so it may be less than perfect, but I've assayed for urea without much difficulty in a relatively simple matrix using an amino column (Shodex NH2P50, I think). I believe my MP was 80/20 ACN/0.1% H3PO4 with detection in the low UV.

Alternately, I would also predict that it would be assayable via reversed phase with ELSD detection if using a perfluorinated ion pairing agent (TFA, HFBA). I've assayed for highly polar urea derivatives that way pretty easily.

CJ
http://the-ghetto-chromatographer.blogspot.com/
5 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry