Agilent 6890 GC analysis of diethylenetriamine
Posted: Wed Mar 12, 2014 6:02 pm
I'm trying to assay diethylenetriamine. I'm using an Agilent DB-35 column (30m x 0.53 mm x 1.0 µm) with a base deactivate single taper liner using base deactivated wool. I've encountered a couple problems that I wanted advice on.
#1 When injecting 100ppm DETA standards, I never seen any peaks until I inject a 1000 ppm standard first. Is it common for this analyte to have to prime the system before low level injections produce peaks?
#2 After injecting a 1000 ppm standard I can achieve 0.999 linearity using 0, 500ppm and 1000 ppm standards and achieve an <10% RSD for 5 injections of the 100 ppm standard, but I'm seeing what appears to be much greater tailing than a comparable chromatogram from a Restek method http://www.restek.com/chromatogram/view ... 7/111-40-0
I don't know how this could be so when all the contacts points ie. liner, column are base deactivated. Where else in the GC flow path could the DETA be sticking to?
Thanks for any ideas.
#1 When injecting 100ppm DETA standards, I never seen any peaks until I inject a 1000 ppm standard first. Is it common for this analyte to have to prime the system before low level injections produce peaks?
#2 After injecting a 1000 ppm standard I can achieve 0.999 linearity using 0, 500ppm and 1000 ppm standards and achieve an <10% RSD for 5 injections of the 100 ppm standard, but I'm seeing what appears to be much greater tailing than a comparable chromatogram from a Restek method http://www.restek.com/chromatogram/view ... 7/111-40-0
I don't know how this could be so when all the contacts points ie. liner, column are base deactivated. Where else in the GC flow path could the DETA be sticking to?
Thanks for any ideas.