Dear Friends,

We have Sykam germany HPLC with uv-vis detector and calrity software. The mobile phase are D.I water and ethanol. We have problem with our HPLC. The baseline is not stable. At detector display the detector signal is 00 (means OK) but in software the baseline is not stable. We have changed the mobile phases with fresh one, checked the column and have found it OK. Please be so kind to suggest us some best solution to make stable the baseline. What factors involve in an unstable baseline? we are new user of HPLC and does not know much about HPLC.

An earliest response in this regard is awaited.

Many thanks & best regards
Khalid
What kind of column are you using, has it been used a lot?
How old is your lamp, how many hours has it been on?
What wavelength are you monitoring?
Thanks,
DR
Image
Turn off the pump, wait a few minutes. Measure the baseline. If more stable then, it's pump-related.
Most important action in this case is viewing and interpreting the baseline.
Show us the problem.
Unstable baseline is many things ;-)

Best Regards
Learn Innovate and Share

Dancho Dikov
Hi,

Thanking you for the kind support.
We are using column C18. and it is almost new. The detector is used since last 3 years. no lamp of the detector has been changed yet. The wavelength is 254nm.

Please let us know if you need more information about our HPLC.

Many thanks

Khalid
Post the plot - baseline so that that we can see it.

The column is absolutely irrelevant in this case.

Best Regards
Learn Innovate and Share

Dancho Dikov
The detector is used since last 3 years. no lamp of the detector has been changed yet.
So, the lamp is a minimum of 3 years old? See if you can find another lamp to try.
Thanks,
DR
Image
Dear Friends,

We have Sykam germany HPLC with uv-vis detector and calrity software. The mobile phase are D.I water and ethanol. We have problem with our HPLC. The baseline is not stable. At detector display the detector signal is 00 (means OK) but in software the baseline is not stable. We have changed the mobile phases with fresh one, checked the column and have found it OK. Please be so kind to suggest us some best solution to make stable the baseline. What factors involve in an unstable baseline? we are new user of HPLC and does not know much about HPLC.

An earliest response in this regard is awaited.

Many thanks & best regards
Khalid
Dear friend,

In the mobile preparation you take precaution of sonicate to degas are not, in the mobile phase preparation, i think your not mix water and ethanol properly & this mobile is take half-an hour for system stabilization purpose.

Thanks and regards,
K.V.Venkata Rao
Change the lamp(s), clean the flowcell(per user manual) and when the system is "spotless" run methanol(with a flow restrictor, low flow) and measure the baseline.You will not have OQ limits but pretty damn close.
Good luck.
9 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry