N2O method on Shimadzu 2014
Posted: Sat Mar 08, 2014 8:22 am
We have a Shimadzu GC-2014 which we've previously used to analyze N2O and CH4 from the soil. It is set up with one 10 port and a 4 port valve. At the moment I'm only interested in the N2O peak. When we use the previously established method (included below) we are able to see the N2O peak, however when we try to do a direct injection we're not seeing the expected peak. For direct injection we are using a combi-pal auto-sampler.
Configuration:
10-port valve - this is where previous samples were injected and fill two 1ml loops.
HayeSep Q (2mm x 6ft)
ECD (N2O analysis) and FID (methane)
Value timing:
0.01min - 10 port valve directs sample into columns for analysis.
0.5min - 10 port valve closes to instrument.
1.45min - 4-port valve opens to detector.
1.85min - 4-port valve closes to detector to vent any acetylene/contaminants
Temperature:
Detector - 300C
Column - 70C with flow of 20 mL min-1
Injector - 200C
Injection:
1.25mL
There are two columns in the N2O line (one before the 4-port valve, and one after) We've recently added the combi-pack auto sampler and have been successful in getting it set up, however when we attempt to do a direct injection the peaks are not stable and one's we're expecting (N2O standard) are not present at all. We've tried adjusting timing and temperatures, but with no luck.
Any help is greatly appreciated and if there is any additional information needed please let me know.
Any help or direction anyone can be provided.
Configuration:
10-port valve - this is where previous samples were injected and fill two 1ml loops.
HayeSep Q (2mm x 6ft)
ECD (N2O analysis) and FID (methane)
Value timing:
0.01min - 10 port valve directs sample into columns for analysis.
0.5min - 10 port valve closes to instrument.
1.45min - 4-port valve opens to detector.
1.85min - 4-port valve closes to detector to vent any acetylene/contaminants
Temperature:
Detector - 300C
Column - 70C with flow of 20 mL min-1
Injector - 200C
Injection:
1.25mL
There are two columns in the N2O line (one before the 4-port valve, and one after) We've recently added the combi-pack auto sampler and have been successful in getting it set up, however when we attempt to do a direct injection the peaks are not stable and one's we're expecting (N2O standard) are not present at all. We've tried adjusting timing and temperatures, but with no luck.
Any help is greatly appreciated and if there is any additional information needed please let me know.
Any help or direction anyone can be provided.