Solvent Peak covers the analyte
Posted: Fri Mar 07, 2014 9:50 pm
by rouzbeh
I am using Hexane as a solvent and my analytes are a mixture of FAMEs.
Everything looks good when injecting in Split injection.
When I do split-less injection the solvent peaks covers my 1st analyte peak.
Re: Solvent Peak covers the analyte
Posted: Sat Mar 08, 2014 9:30 am
by carlos.teixeira
Remain the split mode reducing the split ratio untill you acquire a value that permits aceptable resolution (separation) between the solvent peak and first analyte peak. I wish you got nice elutions!!
Posted: Sat Mar 08, 2014 2:26 pm
by rb6banjo
You only want to go splitless if the sensitivity of your analysis is poor. Splitless should be your first attempt because transfer of the analytes to the column will be more controlled.
Posted: Sun Mar 09, 2014 3:49 pm
by Peter Apps
If you really need to use splitless in order to get the required peak sizes, then try a shorter splitless time.
Posted: Tue Mar 11, 2014 11:51 am
by rouzbeh
Thank you all.i did switch back to split.
Posted: Tue Mar 11, 2014 12:56 pm
by rb6banjo
If you can get the sensitivity you need, your results will be better and you'll be happier.
Posted: Wed Mar 12, 2014 2:49 am
by AICMM
How far out is your first FAME and what is your oven initial temperature?
Best regards,
AICMM