The dynamic mixer included with the system has a volume of 1.2ml!!! Thus, when I run flowrates at 0.2ml/min, it takes nearly 6-7mins for the beginning of the gradient to reach the column... so I'm essentially running an isocratic system when I use a narrowbore column.
To solve the problem, in one instance, I've used a static mixing tee...
http://chromatography.hplcsupply.com/vi ... ing-tees-2
The mixing noise from this tee was nearly 1mAu! In another instance I've used a tee to connect the solvent lines, and a 75ul Perkin Elmer series 200 dynamic mixer. The mixing noise is about 250uAu end-to-end, way higher than I would like it, but better than the mixing tee.

Here at 15min, you can see the moment where I turn on the dynamic mixer and mixing noise decreases.
I have not tried a static mixer cartridge yet, but plan to try a 50-150ul next. Can anyone furnish any advice? Dionex proposes using a dynamic mixer and static mixer in series for best mixing and minimal noise.
Can anyone suggest a low-volume mixer or mixing system for this problem? I'm looking to attain a delay volume of 150uL or less.