Changing display on Agilent pump from bar to psi
Posted: Thu Mar 06, 2014 12:58 am
by ghanson21
I have a variety of Agilent systems in my lab running on Chemstation. I am trying to change the pump display units from bar to psi. Has anyone done this before? The firmware version I am trying to tackle first is B.04.01.
Agilent tech support is looking into this for me but I was wondering if anyone else has already implemented this change?
Re: Changing display on Agilent pump from bar to psi
Posted: Thu Mar 06, 2014 7:13 am
by daniel_aut
Start the Configuration Editor, right-click your installed pump and you should have the option to change pressure units.
Posted: Mon Mar 10, 2014 11:16 pm
by ghanson21
Thank you Daniel_aut....worked like a charm.
I do have some older "gameboy" controllers that I am looking to update the pump units to PSI as well. I have been told that the gameboy units are so old that the display units cannot be changed.
Does anyone know if this is indeed true?
Posted: Tue Mar 11, 2014 6:24 am
by daniel_aut
No, its not true. A machine does only what you tell it to do.