SPECTRA IN NORMAL PHASE HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi all!

Is it normal that PDA-spectra for compounds when working in normal phase are worst than the ones in reversed phase HPLC? :shock:

Is identification of compounds from their spectra still reliable in RP-HPLC?
:o
I noticed that there are some changes between spectra maximums from RP to NP-HPLC (I work with tocopherols & tocotrienols)

thank you in advance
Carmen

I don't know what you mean by "worse".

There is no reason to expect exactly the same spectrum in vastly different chemical environments (aqueous vs. non-aqueous), especially if the chromophores are coupled to ionizable groups.

UV spectra are generally are "information poor". I would not rely on UV spectra for positive identification under any circumstances.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry