Chemstation A.07.01

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
I installed a A.07.01 version of Chemstation on a Win PC. I have a 1315B Agilent Detector.

When I start the CS, the program is not able to identify the Detector. A strange looking icon is shown on the left hand side, however the CS does not know what it is. Even when I add it to the right column, the system is not able to deal with it. I have tried to update the Firmware to 5.08 which should be compatible to this version of Chemstation, however still no success.

Any suggestions ?



Regards
Thats a little bit to few informations.

The system (the DAD) ins connected via LAN to the PC and via CAN to column oven?

Have you installed the bootp application?

When you start your DAD did your bootp work properly (sends the IP to the DAD)?

Can you please make a snapshot of the "strange looking icon"?

Have you added the correct Licence to the chemstation?
Any problem about Agilent GC, HPLC, chemstation, just let us know via engineer201208 at gmail dot com
Thx,
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry