We want to check if we understand it correctly based on what we have read so far:
1. Air and hydrogen gas do not go through the column, but go to detector to ignite the FID.
Helium, as the carrier gas, is the only gas going through the column.
Is that correct?
2. Is yes, when we have helium gas on, and air and hydrogen off, we should be able to measure the helium flow at the detector exhaust vent. But we can not measure any. Does it mean there might be leak in the GC system?
3. We are trying to measure the carrier flow rate using retention time of methane. If we have all gas flows correctly set according to the manual, why is there hexane detected but not methane? What could be the potential cause?
We have manuals and old notes from former students. It is very confusing because the notes from different sources are not consistent, either do the manuals.
Thank you!