Varian 3400 error code.
Posted: Mon Mar 03, 2014 6:56 pm
by zgredxx
Hello. I have some trouble with my unit Varian 3400cx. It's display "Fault 13" and " Fault 37". I don't have manual for my unit. Anybody can tell my what this is errors about? Thanks.
Re: Varian 3400 error code.
Posted: Fri Mar 07, 2014 7:31 pm
by WillyOne
I'm at home now. I will look the manual next monday
Posted: Fri Mar 07, 2014 7:32 pm
by WillyOne
Next monday I will have a look yo the manual.
Regards
Posted: Sun Mar 09, 2014 9:38 am
by marlenl
I can send You the manual in pdf (4,3MB) via email.
13 Power fail occurred (latched)
AC power to the instrument was interrupted, either by a power failure or by turning off the main power switch
37 Power line ground fault (latched)
The polarization of the power line connection is reversed, or the ground connection through the power cord is missing. Normal chromatographic functions are not affected, but the risk of electrical shock is increased for certain maintenance operations and in case of some component failures.
So, first just try to turn 180 the main power plug.
Posted: Mon Nov 14, 2016 3:42 pm
by patsay
Please can you send me (
patsay@mail.ru) the manual for Varian 3400 ?