Tailing on a Polar column

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
Hi,
I am getting tailing on an early non-polar compound. I don't have tailing on later eluting polar compounds and their retention times are not shorter over time. My question is - if the column is degrading do non-polar compounds start to tail ? The column is about a year installed and is used every day so is due for replacement but I don't want to if not necessary. The column is CPWax52CB. Any help gratefully received.
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
When did you last do inlet maintenance ?

Peter
Peter Apps
Hi

Could you post some more details and a chromatogram?

There is a post box near Mornington Crescent :-) :-)

Regards

Ralph

You could add "The Infinite Monkey Cage" to your BBC Radio 4 list :)
Regards

Ralph
Hi,
Thanks for the replies.
I will try to get time to do some proper logical inlet maintenance.
I suspect the pre-packed wool liners we buy in are not always tightly packed enough. I will try to shove some more wool into one and try that (PE PSS injector - 2mm id liner).
Regards
WK

(I will add The Infinite Monkey Cage. Been listening to Rich Hall's US Election show).
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
If you have tailing only of the non-polar component then the most likely culprit is a fragment of septum in the inlet liner. I would be very surprised if packing in more wool would improve anything.

Details on conditions would help.

Peter
Peter Apps
Hi Peter,
I packed the liner with more wool. I swabbed the inlet with methanol. I also cleaned the septum nut hole out and removed septum material in the process.
I checked the column insertion distance in the inlet. One at a time and nothing improved. So I replaced the column and the non-polar component no longer tails.
Using a CPwax52CB it would appear that the peak shape has deteriorated - but the retention times appear to be similar over time. I suppose that is a good thing !
By the way - I have re-installed the liner out of the box - with relatively loose wool.
Thanks for the help
Regards WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
Hi WK

I apologise for deviating from the post :-(

but yes

Rich Hall's US Election show is so good - did you hear the post election one?
Call 555 0101
:-)

Ralph

https://en.wikipedia.org/wiki/Rich_Hall
Regards

Ralph
Hi WK

Thanks for the feedback. Both septum fragments and a deteriorated column can cause non-polar compounds to tail, so if the tailing comes back do the cheap inlet check before you change the expensive column.

Peter
Peter Apps
8 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry