Hi all, our new instrument 5977 cannot pass autotune and the scan seem strange as follow:

Image

The vacuum is OK.

Could anyone help me on this

Thank you
Reposting the picture, it didn't display for me in the Forum.

Image
Time flies like an arrow. Fruit flies like a banana.
What is your peak width? Does the ion focus maximize at something other than the upper limit? Is the repeller voltage something reasonable (<35V)?
Your 502 and 503 look like the same peak same you are not resolving your masses.
What is your peak width? Does the ion focus maximize at something other than the upper limit? Is the repeller voltage something reasonable (<35V)?
Your 502 and 503 look like the same peak same you are not resolving your masses.
Not seeing much 131 at all either, there should be as much of that as there is 69.

The peak widths look quite wide except for 69, or it is just a lot of noise.

I have never looked at a tune from Mass Hunter, which I am guessing this is, but is doesn't look that good either way.

Looks as if the peaks are fronting, which usually means a problem such as dirty quads or a lot of moisture.

Can you do a profile scan of 69, 219 and 502 and post the peak shapes for those masses?
The past is there to guide us into the future, not to dwell in.
Thanks all for inputs.
I have tried all possible. It seem the peaks are not stable, it cannot focus at the exact place.
The CE said, it's due to faulty side board.
We'll wait and see.
Finally, the MSD is OK and it's due to the quadrupole itself, a miss-alignment.
Finally, the MSD is OK and it's due to the quadrupole itself, a miss-alignment.
Wide peaks are normally a sign of problems in the mass filter, glad it was just a misaligned quad and not the side board. :)
The past is there to guide us into the future, not to dwell in.
7 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry