Hello all,
We have several problem with our DSQ II GC-MS which are:
1- The background scan spectrum from 10 to 650 is ok.
2- The air/water spectrum from 10 to 90 is ok.
3- The scan spectrum from 50 to 650 is too noisy and unstable, so we could not check very well the spectrum of the calibration gas due to the big noise.
4- All the diagnostics are ok.
5- The RF Auto tune is passed.
6- All other tunes (detector gain, resolution and mass) are failed.

Please note that we make a full cleaning for Ion source, lenses and the anode, and we make cleaning with dry gas of the prefilter and quadruple. Before the cleaning process, we had a noisy background spectrum, but cleaning solve this issue.

You are kindly requested to give me a solution to fix the noise and make the full auto tune.

Thank you in advanced
You say you have cleaned the prefilter with a dry gas.. I assume you mean you have blow it over with the gas but nothing else?

If so I can highly recommed real cleaning of the prefilter.. I have several times seen this part as the main issue with DSQII systems. Just do NOT sonicate the prefiler.. I have tried that and it cost me a new prefilter..

You write you have poor calgas spectra.. What intensity of the calgas do you have? Could be you have run out of calgas?

I would also double check the assembly of the source... I know this is what everyone always says and I am sure you have done it correctly but I see it time and time again that a slight misallignement of the source parts cause strange issues.
Kind regards
Leadazide
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry