Pressure Increase On Injection - UPLC I-Class
Posted: Thu Feb 27, 2014 4:51 am
I have recently started having a problem with our Waters I-Class UPLC in which the pressure shoots way up upon injection of a sample (as much as 5000 psi at 650 uL/min on top of the normal operating pressure). We used to see small increases in pressure, as would be expected, but this started a few days ago, and I am having a hard time tracing the problem in the instrument.
The pressure jumps as soon as the 6-port rheodyne valve in the autosampler switches into 'Load'. It isn't blocked (injection will still proceed without hitting max pressure at a lower flow rate), and back flushing through the valve doesn't lower the pressure any. I have also checked everything that comes after this valve, and there doesn't appear to be a problem with any of the fittings after the line leading into the column.
Any thoughts would be greatly appreciated.
Thanks.
The pressure jumps as soon as the 6-port rheodyne valve in the autosampler switches into 'Load'. It isn't blocked (injection will still proceed without hitting max pressure at a lower flow rate), and back flushing through the valve doesn't lower the pressure any. I have also checked everything that comes after this valve, and there doesn't appear to be a problem with any of the fittings after the line leading into the column.
Any thoughts would be greatly appreciated.