I am trying to use Empower 3 to run a 6890Plus GC with a 7673 autosampler and a G1512A controller. The 6890Plus GC has dual injectors and FID Detectors but I am only using one side and do not have a column connected to other side. Empower 3 controls the GC and sets the injector, detector and oven temps but it never makes an injection as the GC never gets to a ready state. On the front panel of the GC always get flow, pressure and set temperature errors for the side I do not have a column connected so GC never gets to ready state.
Any suggestions on how to get this configuration to work.

Thanks
As long as it's not ready, there will be no injection made.
There is a calculation going on by using your column ID, length, flow, temperature, pressure, ...
As long as that's not OK, you will get NOT READY.
I guess you need to put on a column, or maybe a plug and turn flow off?

HTH

Ace
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry