Accela pump to PAL autosampler connections

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Does anyone know the pin-outs between the PAL Interface 1 DB15 connector and the Accela pump D-sub25 connector? The manuals lack so much information these days. Nothing states signal in or signal out, TTL or CC, voltage or time.

Also is there a reset for pump stroke count on the Accela?
Hi,

I have just set up an LC-UV system using a PAL and Accela.

The PAL manual has all the pinouts in the back. It can be found here: http://www.palsystem.com/fileadmin/temp ... .pdf&user=

The only connection between the PAL and the pump should be contact closure.

The PAL, Pump, and Detector should all be separately connected to the data system. At least that's how we have it set up and it works great.

Hope this helps!
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry