Manual vs Autosampler Injection Metrohm 881 Compact Pro
Posted: Mon Feb 24, 2014 3:41 pm
I'm am currently using the Metrohm 881 Compact Pro IC fitted with the autosampler. In the last few months we found that there are differences between the peak area obtained using manual injection vs. autosampler injection, despite being injected from the same sample / standard solutions. The difference is in fact more than 50 % (Peak area from Autosampler injection being higher than manual injection). We also found that while the peak area from autosampler injection remained consistent, the peak area from manual injection of similar standard concentrations have reduced by half compared to data obtained 3 months ago. Our standard is sialic acid standard solutions 1.25 - 20 ppm.
We have tested the following:
1) Checked the pump flow. It will flow 10mL of eluent after 10mins with 1mL/min pump rate.
2) Changed the amp body cell, Au electrode and reference electrode.
Recently we have been advised to check the column (Metrosep Carb1-150). But I highly doubt the problem could arise from the column since the difference is observed between the injection methods. We injected full set of standard solutions using both the manual and autosampling and observed similarly consistent differences between manual and autosampling injection peak areas (manual injections' peak areas 50 % lower than autosampling injections' peak areas).
We've been stumped for awhile now and really hope anyone could help advise on the possible source of problem the soonest possible. Thank you.
We have tested the following:
1) Checked the pump flow. It will flow 10mL of eluent after 10mins with 1mL/min pump rate.
2) Changed the amp body cell, Au electrode and reference electrode.
Recently we have been advised to check the column (Metrosep Carb1-150). But I highly doubt the problem could arise from the column since the difference is observed between the injection methods. We injected full set of standard solutions using both the manual and autosampling and observed similarly consistent differences between manual and autosampling injection peak areas (manual injections' peak areas 50 % lower than autosampling injections' peak areas).
We've been stumped for awhile now and really hope anyone could help advise on the possible source of problem the soonest possible. Thank you.