Fatty acid ghost peaks
Posted: Mon Feb 24, 2014 3:59 am
I have been working for a while in bacterial Lipidomics using Thermo Trace DSQ II. All the sudden I started seeing ghost peaks in system blank corresponding to C16:0 and C18:0 fatty acid, underivatized. Since my sample had them, I first gave doubt to the liner after condensation test and changed it. However, it did not solve the problem. I have already changed new inlet house (whole inlet) but the problem is still on. I've installed everything new (liner, graphite seal, sonicated septum holders/ plate, silver seal, and terminal fitting. Tried with new column also and column has no contamination). I was using Merlin septum when I first noticed the problem and the ghost peak intensity was close to 2x10^6. When I tried using a BTO septum (preconditioned) at all new inlet set at 200 C the ghost peak is so worse. It reaches around 10^8. When I use methanol as last syringe cleaning solvent these peaks convert to corresponding FAME on air blank. And, worst part is when I inject BSTFA blank I can see TMS esters of C12:0 to C18:0 fatty acids. I have come up to the tentative conclusion that these septa bleed fatty acids (found a substantiating literature also: http://chromsci.oxfordjournals.org/cont ... l.pdf+html) assuming carrier gas is contamination free (Checked carrier gas too. Condensation test points to inlet). Has anyone encountered similar situation? Share please, I may have been barking the wrong tree........