Newbie question FAME detection GC/MSD
Posted: Sun Feb 23, 2014 4:06 am
I'm a new postdoc fresh out of grad school and joined a lab that has a "new" Agilent 5975c-MSD that has been a giant paper weight for the last two years (well maintained, just not used). I have limited experience with GC - mostly working with analysis of amino acids under the supervision of people with far more experience than I have. Most of my mass spec background is in LC - metabolomics.
We are a lipid lab and I wanted to utilize this instrument to analyze fatty acids. My first attempt yielded mixed results. I ran the Supelco FAME 37 ( in methylene chloride) which I diluted in hexane in the following concentrations 1mg/ml, 100ug/ml, 10 ug/ml, 1ug/ml and 100 ng/ml.
I ran the method that was already loaded for FAME's
Column: HP-5MS 30 m x 0.25 mm x 0.25 μm
Injection: 1 ul
Method: 50oC 1 min, 25oC to 175oC, 4oC/min 235oC
Split: 1/50
Carrier: He
Two problems:
1) My chromatogram only had 28 peaks corresponding to 24 FAMEs. Most of the longer chain FAMEs were missing >20:0 and the 18:0 was also missing.
2) I only had peaks from the sample at 1mg/ml. I ran the samples in order of lowest concentration to highest - so this would have been the last vial to run in the series. Total volume was 100 ul in a vial with 100 ul insert.
I am a little lost, I am used to LC where I can tweak the method a bit and get better results. I tried a few tweaks, but nothing seamed to work. Any suggestions?
Injection: 1 ul
Split: 1/50
Carrier: He
Two problems: