Empower 3 and 6890Plus GC

Posted: Fri Feb 21, 2014 9:14 pm
by margurn
I am trying to use Empower 3 to run a 6890Plus GC with a 7673 autosampler and a G1512A controller. The 6890Plus GC has dual injectors and FID Detectors but I am only using one side and do not have a column connected to other side. Empower 3 controls the GC and sets the injector, detector and oven temps but it never makes an injection as the GC never gets to a ready state. On the front panel of the GC always get flow, pressure and set temperature errors for the side I do not have a column connected so GC never gets to ready state.
Any suggestions on how to get this configuration to work.

Thanks

Re: Empower 3 and 6890Plus GC

Posted: Mon Feb 24, 2014 9:53 am
by aceto_81
As long as it's not ready, there will be no injection made.
There is a calculation going on by using your column ID, length, flow, temperature, pressure, ...
As long as that's not OK, you will get NOT READY.
I guess you need to put on a column, or maybe a plug and turn flow off?

HTH

Ace