Agilent 1200 mwd problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
New to hplc with no help. I am running an Agilent 1200 with a mwd. The deut. lamp died so I replaced the lamp. I had good peaks, left for the night and when I came in in the morning, the peaks have disappeared. I am thinking it is a connection problem because I still have peaks in my fluores. detector. Has anyone seen this or know what can be replaced?
Power down the detector and then turn it back on to let it run its self-test. If the new lamp has failed, it should fail the POST with an error code that should give you some information as to the problem. If it passes, make sure that the controller is talking to the detector (again, there should be a specific error message if it isn't). Assuming all that checks out OK, look to see if the wavelength setting is correct.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 37 users online :: 1 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 36 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry