Empower 2 Cuvette Reading Help

Posted: Tue Feb 18, 2014 8:24 pm
by tristanewalters
Hello all-

We are trying to qualify a Waters Alliance instrument with a 2998 PDA detector. We are wanting to do a single PDA scan of a NIST holmium oxide standard. Can anyone provide the steps we need to acquire the sample through Empower 2? I am having a brain fart on how to do it. Thanks!

Re: Empower 2 Cuvette Reading Help

Posted: Wed Feb 19, 2014 12:47 pm
by shaun78
Make a 0.0 uL injection of a sample from within Empower using your desired sample set for your desired length of time.

You will probably want to use a PDA scan of the "sample" for about a minute or so.