saturn - Pump Error: Controller fault occured

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi

In our Varian saturn 2000 GC-MS. I am getting error "Pump Error: Controller fault occured". Our Foreline pump is ok this is related to the diffusion pump. While starting diffusion pump it failes and report "Heater overtemp". The detail message it throws -

"The diffusion pump is reporting a fault: Diffusion pump heater has overheated. if this was true on entry to diagnostics. This fault was reset but is still present. Please correct the problem and try again after the heater has cooled."

I have turned off and allowed to cool and turned on again but it still throw the message. Do I need to reset somehow. Diffusion pump seems not starting up. Any direction to troubleshoot will be appreciated. Thank you
I figured the problem. It was the thermocouple attached to diffusion pump was not resetting back to close the circuit. I took it out and tapped couple of times which reset it. Also checked conductivity between the terminal via voltmeter to make sure.
2 posts Page 1 of 1

Who is online

In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 30 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry