Thermo Fisher DSQ II gc/ms Autotune fails

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Hi to everyone,

i'm having some trouble when running the autotune of my DSQ II GC/MS. When I run the default autotune (RF frequency calibration, detector gain calibration, resolution calibration (positive ions), full automatic tune and leack check), the system fails to pass the detector gain calibration (error: cannot perform gain calibration). I tried to unflag the gain calibration, but it still fails to pass the autotune (the message says "creating tune report for positive ions" and "error: cannot calibrate resolution). I tried to find some answer on the Thermo Fisher manuals, but they don't say much about autotune failures. I read a previous discussion here that said the problem could be in the lens voltage in the MS, but the readings from the instruments barely differs from the set values. Can some explain me what is wrong with my GC/MS?? Thanks a lot!!
have you tried cleening the system?
Problem might be caused by heater PCB and/or lenses that are dirty
Did you run the multiplier test in Diagnostics?
I had this issue for months! Run diagnostics in your tune software. If they pass, try the following:

Run air/water, background and FC43. If you don't see anything check (top up?) your cal gas, rinse and repeat
Clean your dirty ion source
Up your detector gain (but not too much, I think it shouldn't ever go above 2000??)
Still nothing? Check your filament's glowing (just take the top off)
Still bad? Maybe it's your S/N ratio? Replace / cut your column

Oh and call thermo, they can help you with diagnostics over the phone. I kept sending them screen caps of my complete lack of signals.
Hi,

for me it failed in this phase because I had a bad contact with Lens 1. After correcting that i still got calibration failed because i tried to do it too early. Even if it shows 50mTorr, you should let it rest for a couple of hours after a cleanup.

Regards
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry