May I use more aqueous phase than organic phase for reverse phase HPLC column, such as C18 column?
like 60% aqueous phase, 40% organic phase?


I am an newer for HPLC. Hope it is not silly question.

Thanks,
Sophie
May I use more aqueous phase than organic phase for reverse phase HPLC column, such as C18 column?
like 60% aqueous phase, 40% organic phase? I am an newer for HPLC. Hope it is not silly question.
Thanks, Sophie
Yes. We go low organic-high aqueous frequently as starting point for gradient elutions for polar species.

There are some C18 columns that cannot well tolerate less than 5% organic, so to go that low, check with supplier.
For a standard RP column I would recommend to follow the 5% rule. Keep always minimum 5% aqueous phase or on the other side 5% organic phase.
Gerhard Kratz, Kratz_Gerhard@web.de
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry