sepearation of Lactate and acetate on PRP-X 100 column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I would like to detect Lactate, Acetate and sulfate using PRP-X 100 column with indirect UV method. The problem is, I am not able to get separation of lactate and acetate. (mobilephase 4mM p-hydroxy benzoaic acid pH 10) I need some help in this regard.

thanks in advance
Try decreasing the pH.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
There is a pretty old paper:

The determination of lactate, acetate, chloride and phosphate in an intravenous solution by non-suppressed ion chromatography; Haddad,P.R. Croft,M.Y.; Chromatographia, 1986 / 21 / 11, P. 648 -650.

Eluent: 0.3 mM phthalate (pH=6.0), 30% ACN

Acetate and lactate are well separated within 9 min. But no information about sulfate. Iassume that ic comes pretty late.
The paper shows as well that acetate/lactate separation strongly improves from co-elution at 0% to approx. 4 min. RT difference at 50 % ACN.

You might try ACN with your eluent as well, may be with pH change as well.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Use a lower concentration of phthalate(1 mM) and add some MeCN (5-15%).

Bear in mind that conditions that facilitate separation of these early eluting organic acids will likely lead to excessively long retention times for sulfate. You may consider a step gradient to 10 or so mM phthalate once acetate and lactate elute in order to get sulfate off in a reasonable time.
Thank you very much for nice suggestion, will try and get back to you.

Thank again
Separation of lactate and acetate by HPLC :
- mobile phase : 1000 ml distilled water containing 1.5 ml phosphoric acid .
- column : 150 × 4.6 mm , C18
- detector : UV 210 nm
- injection volume : 20 micro litre
- column temp : 25
- flow rate : 1 ml/min
I am applying this simple method usually and it gives very good results

ph/Amr Tarek
Instrumental Unit Head at Otsuka pharmaceuticals
Thank you very much for nice suggestion. will try one by one and get back to forum if get some success.

Hiren
7 posts Page 1 of 1

Who is online

In total there are 163 users online :: 0 registered, 0 hidden and 163 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 163 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry