LC-MS/MS method for polar compounds

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hi All,
I got 3 analytes (no structures provided) with little information. Compounds similar to Cytidine monophosphate and one unphosphorylated. Trying to get a LC-MS method for quantitation in biological matrices. Due to polar nature of the compounds, there is no retention of phosphorylated analytes on a regular C18 columns. Do I need to go with ion-pairing chromatography, which I'm trying to avoid if at all possible. I believe the monophosphate is polar acidic compound. Is HILIC an option here? Any suggestions please - Thanks!
bisphosphonate,monophosphonate are not ionise in MS/MS due to their High polarity.eg.Risedronate.
So we should go for derivatives method.
After derivative their molecular weights and properties are chanhed and ionise in MS/MS.
bisphosphonate,monophosphonate are not ionise in MS/MS due to their High polarity.eg.Risedronate.
So we should go for derivatives method.
After derivative their molecular weights and properties are chanhed and ionise in MS/MS.


This is not correct; phosphonates are detectable by ESI-MS in the negative ionization mode.
Hi All,
I got 3 analytes (no structures provided) with little information. Compounds similar to Cytidine monophosphate and one unphosphorylated. Trying to get a LC-MS method for quantitation in biological matrices. Due to polar nature of the compounds, there is no retention of phosphorylated analytes on a regular C18 columns. Do I need to go with ion-pairing chromatography, which I'm trying to avoid if at all possible. I believe the monophosphate is polar acidic compound. Is HILIC an option here? Any suggestions please - Thanks!
HILIC might help you. There are some papers/application note that might help you :
  • http://www.ncbi.nlm.nih.gov/pubmed/22816250
  • http://www.merckmillipore.com/showbrochure/201008.054.ProNet.pdf
From my experience you can separate phosphonucleotides on a C18 column (AMP/ADP/ATP) with just isocratic flow rate with say, phosphate buffer.
It might not be possible to accurately analyze those compounds in their pure form (without derivatization). The reason is that they are highly polar and they get accumulated on stainless steel (you can find many articles about this topic). Almost everyone neglects this fact.

Good Luck,
Dan
6 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry