quercetin and its glycosides are amongst the most common and studied compounds in the plant kingdom, so there are certainly hundreds of published methods, possibly thousands. Every year there's another paper saying "we put some flavonoids down a C18 column and look! They all came out separate!"
The fundamental problem is that a crude extract of tomato contains a phenomenally large number of compounds that will absorb in the UV, so you will never get them all separated in a single LC run. If you believe that something like rutin is the most abundant compound in your mixture, then you may be able to come up with an LC run where the rutin peak is sufficiently separated from other biggish peaks, and contaminated only by tiny peaks, such that you trust the result, but HPLC is always going to be wobbly on crude extracts.
But James_Ball is right; the more resolution you have, the better your chances, so it may help a bit to change to a Kinetex-type column (solid core) so you can get the effect of smaller particles/longer column while keeping a back-pressure that's appropriate for your system. You can experiment with flattening the gradient over the region where the coeluting peaks appear (but this has diminishing returns; once you're flat enough, you get little by going flatter). If you can't resolve the unwanted peak from genuine quercetin, then ultimately a different mechanism of retention is good. James_Ball suggested CN, I would add something like phenyl-hexyl to the list, or if you're a Phenomenex user, PolarRP. I haven't tried the perfluorinated phenyl columns myself, but have heard amazing things about them. The point of a phenolic column (and is CN doing the same thing, James??) is that it will bind by pi-pi interactions, so it's good for phenolics, which quercetin is, but selectivity is very different to C18. But note: these columns need to be used with methanol rather than ACN as the pi electrons in ACN compete.
Personally, on flavonoids, I like to have MS as a back-up, but I know that's a luxury not everyone has. My feeling is that UV is more quantitative, and gets round the problem of standards for things that aren't commercially available, but if my UV peak areas don't correlate with MS peak areas, then my UV detector has been misled by a coeluting absorbing compound.