HPLC
Posted: Fri Jan 24, 2014 5:48 pm
by makula123
we have Agilent 1200 Binary Pump. How to get rid of air in the solvent line?
Re: HPLC
Posted: Sat Jan 25, 2014 5:05 pm
by tom jupille
Read the user manual. The procedure is there.
Re: HPLC
Posted: Tue Jan 28, 2014 7:21 pm
by James_Ball
If you do the normal purge and still see bubbles in the solvent line, then the bottle filter frit could be clogged, which will cause bubbles to form from cavitation in the solvent line. If the bubbles are worse the faster you run the pump this is probably the problem.
Re: HPLC
Posted: Mon Feb 03, 2014 6:28 pm
by ph/Amr Tarek
Follow this common procedure to remove any bubbles in the lines :
1) immerse the two lines in a reservoir containing 100 % methanol .
2) open the pump purge .
3) adjust flow rate 2ml/min .
4) wash the two lines at the same time ( 50% : 50% ) with 100 % methanol for about 30 min .
ph/Amr Tarek