Use HP Flow Calculator software
http://www.chem.agilent.com/cag/servsup ... s/GCFC.htm to determine what head pressure you need to get your desired flow rate through the capillary (enter in your dimensions, temperature, etc.).
Then temporarily open up the total flow valve , maybe halfway. Then set the column head pressure to the desired psi. Now, you know the column flow from the calculator since you know the head pressure and temperature. Measure the flow through the purge vent to get split flow, can calculate split ratio from column flow and the total of column + purge vent flows; adjust the purge flow using the total flow knob until it is at the dired flow rate (adjust the head pressure knob a little if necessary to maintain the desired head pressure). Obviously if the total flow valve is set too low, there will not be sufficient carrier gas available, and if it's too high, then the purge flow will be higher.
"I have three valves to adust: column head preassure, total flow and purge valve, and a bubble flowmeter. I need a flow of 20ml/min and a split of 1:20. Please help!!!!!"