As a foreword, I don't have any pratical experience with Waters UPLCs, neither the "classic" UPLC nor the newer H-class or I-class. My experience with UHPLC is based on hardware from other companies. Nevertheless, I'd like to share some thoughts.
- Concerning the transition HPLC -> UHPLC, I've also made the experience that routine QC staff often seems to shy away from UHPLC. It's new, it's unknown, it's expensive, the pressure is so high, the column is so tiny, whatever. They rather use their wellknown 250x4.6 monster with a sequence that runs three days than having the same work done with a UHPLC method in three hours. R&D staff, on the other side, tends to embrace UHPLCs and the possibilities they offer. I know, generalizations like this are always problematic, but I think there's some truth in this one

.
- If you want to use classic HPLC methods on UHPLC, no matter which manufacturer, be warned that you might run into problems. Lower dwell volumes lead to different gradient shapes, lower system volumes may lead to peak shape problems when using "weird" solvent/mobile phase combinations, detectors can be less sensitive because of lower volumes, ... UHPLCs are not worse than HPLCs for running those methods, they are just different. What it boils down to is that you will see how robust your HPLC method really is

.
- Generally, what I really dislike about the whole UHPLC hype is that everyone behaves as if the wheel has been reinvented. It's still HPLC, even if you put an "U" at the beginning

. The same old-school rules about chromatographic hygiene apply and some common sense is still needed. It's nothing new that buffers might precipitate if you're not flushing your columns and your system before shutting down. If you properly learned what you're doing and follow the rules, I don't think that you will have much more problems with UHPLCs than with HPLCs. We don't use "UHPLC-grade special filtered Acetonitrile" or similar stuff, we usually don't even filter our buffers, and we don't have any problems with blocked capillaries. An inline filter still is your best friend

.
- Concerning solid-core vs. sub-2-micron, one should be very careful not to mix up efficiency and selectivity. I think it's rather common sense now that 2.6-2.7µm solid-core particles can deliver about the same efficiency as sub-2-micron fully porous particles. This does NOT mean that any solid-core column will give you the same separation as a sub-2-micron column. I'd say that most examples in column manufacturer brochures that are meant to prove that fully pourous is better than solid-core or vice-versa (depending on what the manufacturer is selling

) actually don't show better general performance of the respective column in terms of efficiency but just a slightly different selectivity. Which means that column A is not better than column B but column A is only better suited for THIS particular analysis.