Detect tripropylene glycol
Posted: Wed Aug 17, 2005 10:16 am
by laucchiun
I have one tripropylene glycol sample to run for assay. My column is supercolwax-10, 60m,0.32mmID, 0.5um. But seems that dipropylene glycol and tetrapropylene glycol cannot separate from tripropylene glycol. How?
Posted: Wed Aug 17, 2005 3:13 pm
by Consumer Products Guy
In my lab we'd try first to dissolve in DMF, derivatize with BSTFA, and use a non-polar capillary column. Expect multiple peaks.
Equity 1
Posted: Wed Aug 17, 2005 4:24 pm
by chromatographer1
On a 30meter Equity 1 0.53mm 1.5µm film splitless with pulsed injection, oven @ 50°C to 200°C @ 8°/min gives 4 peaks for diproylene glycol at about 12 min retention for a 100 ppm solution.
This is without derivitization but with excellent peak shape.
I would prefer C P Guy's method.
I suspect that you may not be eluting the other analytes from the column or that your injector temperature is too low.
Try using a 5 or 10m column, preferably a 1.0µm methyl silicone or a thin film FFAP column (instead of a Wax) if you need an analysis on a polar phase.
But I would follow C P Guy's method on a 0.5µm 0.32 or 0.25 mm ID MS non-polar column with derivitization as a trimethylsilyl ether derivative .
You could also make the di acetate derivative of the diols and run on non-polar or polar as well.