I want to develop a GC analysis method for some halogen gases. the column I used is PC- silicaPOLT, the peak tailling is very serious. Is there anybody who met the same problem as me. I'll appreciate it if you can share your apinion with me.
-
- Posts: 34
- Joined: Thu Jul 28, 2005 8:06 pm
Can you give a little more detail? Such as the solvent used?
Thanks and Best Wishes
-
- Posts: 5
- Joined: Wed Jul 20, 2005 10:24 pm
actually the samples I need to analyze are gas, the carrier gas I used is hydrogen and the detector is TCD (thermal conductivity detector). I do not know whether the column I choose is correct? or The column is ok, just need to optimize some parameters such as flow rate of the carrier gas, the oven temperature and the volumn size of the sampling loop. actually I have tried what I can do. for example, change the sampling loop size, dilute the sample, optimize the flow rate of the capillary colum, optiamize the ramp rate of the programmed temperature of the oven.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You should try to use a porous polymer packed or capillary column.
Alumina plot capillaries have been used extensively. Silica plot capillary may not have the proper pore opening for the analytes you are measuring. That is probably the cause of the tailing. You need a larger pore silica.
You could also try a cyclodextrin capillary.
Rod
Alumina plot capillaries have been used extensively. Silica plot capillary may not have the proper pore opening for the analytes you are measuring. That is probably the cause of the tailing. You need a larger pore silica.
You could also try a cyclodextrin capillary.
Rod
-
- Posts: 5
- Joined: Wed Jul 20, 2005 10:24 pm
thank you very much. I really appreciate it.
-
- Posts: 34
- Joined: Thu Jul 28, 2005 8:06 pm
It is strongly recommanded not to use TCD for active compounds since it has the potential to destroy the TCD filament. Most halogen compounds are considered active.actually the samples I need to analyze are gas, the carrier gas I used is hydrogen and the detector is TCD (thermal conductivity detector). I do not know whether the column I choose is correct? or The column is ok, just need to optimize some parameters such as flow rate of the carrier gas, the oven temperature and the volumn size of the sampling loop. actually I have tried what I can do. for example, change the sampling loop size, dilute the sample, optimize the flow rate of the capillary colum, optiamize the ramp rate of the programmed temperature of the oven.
the tailing might be due to the analyte residue in TCD. You might want to switch your detector to ECD for halogen compounds.
Thanks and Best Wishes
6 posts
Page 1 of 1
Who is online
In total there are 55 users online :: 0 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 55 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 55 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science