We are using Agilent's 6890 GC with ChemStation B.03.02[341] with an FID detector only. When a sample runs, a solvent peak dominates the chromatogram within the first few minutes, and we would like to know how to keep the solvent signal from being acquired in our data. Thank you.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Inhibit integration until after the solvent has eluted.
Peter
Peter
Peter Apps
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Agree w/ ^ but wish to add a bit...
One should pay attention to the solvent front as it can be a less than subtle indicator of a leak in the column compartment. You need not include it in your data, but it should be inspected periodically. I might be more inclined to inhibit integration as suggested above and also manually scale chromatograms so that it goes off scale or is not displayed as a result of starting your x axis after it has eluted. This way, one could go back and look at it if a leak, flashback or other problem is suspected.
One should pay attention to the solvent front as it can be a less than subtle indicator of a leak in the column compartment. You need not include it in your data, but it should be inspected periodically. I might be more inclined to inhibit integration as suggested above and also manually scale chromatograms so that it goes off scale or is not displayed as a result of starting your x axis after it has eluted. This way, one could go back and look at it if a leak, flashback or other problem is suspected.
Thanks,
DR

DR

-
- Posts: 17
- Joined: Tue Dec 10, 2013 8:00 pm
I agree with the last two posts but I would also like to add something
You can increase your flow to lessen the solvent peak but you will also decrease you retention times and possibly decrease you recovery.
The solvent peak will also tell you about your systems heath. Fat peak could show you that you have a dirty glass liner or a plug in your column
Jones
You can increase your flow to lessen the solvent peak but you will also decrease you retention times and possibly decrease you recovery.
The solvent peak will also tell you about your systems heath. Fat peak could show you that you have a dirty glass liner or a plug in your column
Jones
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
And, if you have a choice of solvent -- carbon disulfide is a wonderful solvent for FID. Do be aware that it will react with some compounds, like amines - so it is not appropriate for some samples.
5 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science