System with 2 injectors and detectors.
Discussions about GC and other "gas phase" separation techniques.
			6 posts
							Page 1 of 1
					
	
						Do you need to plug the detector and injector that is not used? Is it possible to cause problems by running the system with the injector and detector not plugged? I have ordered a couple of nuts, but I currently don't have them in hand and I need to run a few samples using the one injector and detector only. I am just interested to know if I will put the system in danger by having so much hot air entering the injector and detector not in use.
					
									- 
																									
																							 - CE Instruments
 
						The answer with most GCs is no.  You should have independent control of the gases to the detector and injector ?  Assuming you have independent control just switch off the spare injector and detector along with their gases.  If for some reason your GC does not independently control the gases you can always just run and old column between the spare injector and detector ?
					
									- 
								
 - Posts: 5433
 - Joined: Thu Oct 13, 2005 2:29 pm
 
						If one detector is an ECD it is a good idea to keep the make-up gas flowing, which will need the column connection to be plugged. You do not need a special nut for this - just use an ordinary column connector with a short piece of wire instead of the column.
Peter
									Peter
Peter Apps
				- 
								
 - Posts: 350
 - Joined: Tue Jun 28, 2011 7:30 pm
 
						My habit is to keep the extra inlet and detector plugged with a minimal (1 to 5 cc/min) amount of carrier or make up gas flowing to prevent oxidation.
					
									- 
								
 - Posts: 154
 - Joined: Tue Jan 26, 2010 11:08 am
 
						great advice. I think that since the neighbouring detector and injector always heat up, it would be a good idea to have inert gas flowing all the time.
					
									- 
								
 - Posts: 3477
 - Joined: Mon Jan 07, 2013 8:54 pm
 
						If it is setup for capillary columns just put a short piece of guard column between them with a low flow, probably the easiest fix you can do(or just a short piece of column similar to what is installed in the other position).
					
									The past is there to guide us into the future, not to dwell in.
				
			6 posts
							Page 1 of 1
					
	Who is online
				In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 