Waste analysis

Posted: Tue Aug 16, 2005 2:54 pm
by ozric
Hi all. I am currently working in the waste business using GC-FID(mainly organic solvents from phamachem industry). I need a "catch-all" method for anlysis of mixed solvent wastes. All the usual suspects ie EtOh, MeOh, IPA, DCM, ETAC etc... Any suggestions??

Posted: Tue Aug 16, 2005 8:04 pm
by Mark
Ozric,

Possibly a 3 - 5 micron 30 m x 0.53 mm DB-624 column may give you all of the alanytes you need. We use one for general volatiles work with good results. Programmed from 40C (hold 3 mionutes) to about 200C should give you an excellent volatiles range.

Regards,
Mark

Waste analysis

Posted: Wed Aug 17, 2005 8:02 am
by ozric
Thanks Mark

I have the system running and identifying all components. What I need to do is quantify each component, I'm having trouble coming up with suitable standards!! :?

Posted: Wed Aug 17, 2005 6:53 pm
by Mark
ozric,

If you are doing these by FID then you have to have the individual standards because of the large range in response factors from these various analytes. I don't really know of a practical way around this.

Regards,
Mark