hydrogen analysis

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I want to set up an analysis method for Hydrogen with the Agilent 6890GC with the TCD as the detector. is that possible? if it is true what kind of column I need to choose? as for the carrier gas which one is better, helium or nitrogen? thanks
You should use nitrogen carrier gas if you want linearity of response over a wide range.

Do not use Helium unless your measurement is starting at about 10% and above. The measurement will not be linear at the low end or accurate below 10% unless it is below about 4% to 0%.

You can use a packed Carboxen 1000, or Carboxen 1010 capillary, or packed or capillary molecular sieve columns, 5A or 13X.

Any questions, feel free to contact me directly.

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rgeorge@sial.com

thanks
3 posts Page 1 of 1

Who is online

In total there are 102 users online :: 0 registered, 0 hidden and 102 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 102 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry