Hi! We have an Ailent Technologies 5975B inert MSD connected to Agilent 6890 Network GC System. Upon, boot up, a message would appear: "No response, cannot bring MS on line(possible bad address or existing lock [error 204]". I think it has something to do with the communication (probably) of the software and the gc system. Any thoughts on this? Thanks!
-cindy
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Your computer is not talking to your mass spec. Reboot the mass spec and see if it will communicate. You shouldn't have to vent.
If that doesn't work you may have a board problem. I have a 5973 that has done this twice, the second time it required a smart card replacement.
If that doesn't work you may have a board problem. I have a 5973 that has done this twice, the second time it required a smart card replacement.
-
- Posts: 129
- Joined: Wed Apr 01, 2009 4:02 pm
Have the bootp program open so that you can watch and see if the MSD IP comes up.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
You may also want to re boot your ethernet switch.
-
- Posts: 11
- Joined: Mon Sep 19, 2011 10:24 am
Thank you very much for your replies.
We reboot the MS without venting and everything got back to normal again.
Cheers!
We reboot the MS without venting and everything got back to normal again.
Cheers!
-
- Posts: 1
- Joined: Sat Apr 14, 2012 12:38 am
I have the same problem, but I can not open ChemStation. We check the network cables, communication between the ms and gc with the computer. Try to repair the software, then reinstall the software and still does not open the program. Help please.
-
- Posts: 11
- Joined: Mon Sep 19, 2011 10:24 am
Hi Laura!
What do you mean when you say you can not open the ChemStation?
What do you mean when you say you can not open the ChemStation?
-
- Posts: 1
- Joined: Wed Nov 21, 2012 2:45 pm
I have the same problem and also Chemstation can't be opened .... I don't know how to restar the MS without venting... somebody can help me?
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
You can turn the MS off for a second or two without venting. just hit the power switch. wait a few seconds and turn it back on. What do you think happens in a brown out?
If you have a 5975 and a 6890N there should be no need for BootP. That was for older instruments with jetdirect cards. The 5975 and 6890N are full network capable instruments.
If you have a 5975 and a 6890N there should be no need for BootP. That was for older instruments with jetdirect cards. The 5975 and 6890N are full network capable instruments.
-
- Posts: 2
- Joined: Thu Jun 21, 2012 6:44 pm
I just had the same error message on my GC-MS this morning. After re-booting the Ethernet switch, it went normal again.
10 posts
Page 1 of 1
Who is online
In total there are 179 users online :: 0 registered, 0 hidden and 179 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 179 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 179 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science