Agilent 1100 modules
Posted: Tue Jan 14, 2014 5:21 pm
by KM-USA
I hope this is an easy question for some here. I have a manufacturing QC site that has an Agilent 1100 system and uses a game-boy type controller. So the question is: do ALL Agilent 1100 modules have built-in GPIB ?
Because we have unused licensed copies of Chemstation and GPIB boards/cables here and could readily move them to ChemStation if the modules already contain those. Their 1100 modules are quaternary pump, column compartment, autosampler, degasser, and VWD.
Thanks.
Re: Agilent 1100 modules
Posted: Thu Jan 16, 2014 8:46 am
by asm
The degasser don't have GPIB, the rest have.
But - why would you use GPIB when you can use LAN?
As far as I know you only need to connect one modul per GPIB to the PC. (Normally it's the detector.)
The rest of the modules are connected via CAN.
Posted: Thu Jan 16, 2014 8:16 pm
by nimbi63
The 1100 LC had GPIB built in, with a slot to plug in a LAN card. The LAN is much faster, but make sure you don't plug a LAN cable into the CAN bus connectors as they both used the same connector type.
The GPIB also requires the Agilent I/O Libraries be install and configured.
Posted: Sat Jan 18, 2014 5:07 pm
by Consumer Products Guy
Thanks.
Ask them to shoot and send you good digital photos of the rear of the modules.