Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi,

we have a number of HPLC systems in the lab here.

In particular i have a PDA detector Code SPD-M10Avp

My question is what is the Tungsten lamp specifications for this system?

What is the voltage and wattage required in this bulb?

It does not say on the lamp taken out of the system, and is not in the manual.

Thank you very much.

Regards

John
It is a 12 volts / 20 Watt tungsten lamp
Thank You

Regards.

John