Empower report issues

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
OK....I am having some issues with Empower 2 reports.

The reports I am generating include the concentration calculated by the software. For the standards, the concentrations are reporting fine. However, in the same reports, the controls and unknowns have concentrations that are reporting as 10 fold less than they should (but...the standard deviations are calculating as they should based on the correct concentration).

Example--

Control Value--20.0 ug/ml
Concentration is reporting as- 2.064 ug/ml (should actually be 20.64)
% Deviation is still reporting as should- 3.20%

Any ideas on how to fix this?
Check your processing method - is it set for conc. or amount? If amount, do you have the right dilution factor for samples?
Thanks,
DR
Image
After checking it appears that one setting was for concentration...one for amount. After I changed both to concentration it fixed the problem!!! Thank you sooo much...I never would have looked at that and this has been driving me crazy for 6 months!

So thank you!!! :P
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry