we have some problems identifying a component from a full scan chromatogram.
It is the small peak at 13.40 in the first chromatogram.
Using the NIST it turned out to be anthracene or phenanthrene with a match of over 900
It is not present in the blank (second chromatogram).
In order to confirm its ID I've injected a PAH-mix (3th chromatogram). (Peak shape is bad because the concentration is to high.)
the peaks at 14.22 and 14.39 are anthracene and phenanthrene but these are +/- 1 minute later than our unknown that has been identified as anthracene by NIST.
to make sure there wasn't a problem with flow etc I've made a mix of both the unknown and the PAH-mix and this produced chromatogram 4.
You can clearly see al 3 peaks (the peak at 13.96 is irrelevant to this problem) and they have stayed at their retention time.
comparing the mass-spectra of peak 13.40 and 14.22 gives the spectra below. They are identical !!!
so the question is:
can the small peak at 13.40 be Anthracene
if not : what is it ??