Page 1 of 1

shoulder peak

Posted: Mon Aug 15, 2005 11:07 am
by moon704520
I have been using the kromasil columns for a long time , I have been using the TEAP pH3.0 as the mobile phase,gradully i find that after a period of time shoulder peak will appear by using kromasil c-8 column ,yet this does not happend on the kromasil c-8,can anyone tell me
why is so ?

Re: shoulder peak

Posted: Tue Aug 16, 2005 3:52 am
by tom jupille
I have been using the kromasil columns for a long time , I have been using the TEAP pH3.0 as the mobile phase,gradully i find that after a period of time shoulder peak will appear by using kromasil c-8 column ,yet this does not happend on the kromasil c-8,can anyone tell me
why is so ?
As far as I know, kromasil c-8 should behave just like kromasil c-8. :wink:

If you really meant that the c-8 column is less stable than the c-18 column, then this is fairly typical. The shorter-chain bonded phase is more likely to be lost with low-pH cleavage of the silyl-ether linkage.