James,One thing I did do was convert the Electron Multiplier to the ETP one from the K&M version which solved a problem with non-linear responses which increased with increasing calibration standard concentrations( mostly a problem with volatile purge and trap analysis ).
Thanks for that info Aldehyde. I was unaware of the new MS. It looks great! However, it appears to me that they are doing away with Chemstation. They don't mention it as a software for the new 5977. They only show Masshunter and OpenLab.Agilent has a new MS coming out soon that apparently has a 2-3x sensitivity/detectability boost due to a change in the ionization chamber--the 5977A MSD. A voltage is applied to the drawout plate which focuses more ions through the source body than before, usually they just bounce off the plate and hang around in the ionization chamber.
http://www.chem.agilent.com/en-US/promo ... solve.aspx
We switched all of our 5973 and 5975 systems over to the ETP and even the semi volatile analysts said it improved the linearity. I was floored the first time I used one when the autotune tuned at around 800V. I have yet to have one just die like some of the other types have before, they just slowly lose sensitivity over time though most I have run for at least three years before replacing them. ETP even says you can try cleaning them with solvent if they become contaminated with high boiling organics but I haven't tried that yet. In the volatiles lab I calibrate from 0.5 to 200 sometimes 400ppb with no problems usually using a 0.25id column, could probably extend that with our 0.18 columns. Had one unit testing using Hydrogen carrier and the multiplier seemed to work fine with that too.James,One thing I did do was convert the Electron Multiplier to the ETP one from the K&M version which solved a problem with non-linear responses which increased with increasing calibration standard concentrations( mostly a problem with volatile purge and trap analysis ).
Do you like the ETP multiplier? I bought one but I haven't installed it yet. We are hoping to increase sensitivity (just a little bit) for sub-ppb PCB analysis. We also calibrate over almost 5 orders of magnitute right now (0.5 to 2000 ppb).
Acquisition is with Masshunter, however the interface is designed to look like MSD Productivity Chemstation so there is not a whole lot of difference there.Thanks for that info Aldehyde. I was unaware of the new MS. It looks great! However, it appears to me that they are doing away with Chemstation. They don't mention it as a software for the new 5977. They only show Masshunter and OpenLab.Agilent has a new MS coming out soon that apparently has a 2-3x sensitivity/detectability boost due to a change in the ionization chamber--the 5977A MSD. A voltage is applied to the drawout plate which focuses more ions through the source body than before, usually they just bounce off the plate and hang around in the ionization chamber.
http://www.chem.agilent.com/en-US/promo ... solve.aspx
Hi! Have you purchased your new GCMS? What did you get? Thanks!We are currently in the process of purchasing a new GC/MS system. We have decided that our top three systems are Thermo Sci (Trace 1310 with ISQ MS), Shimadzu (GCMS QP2010 Ultra), and Bruker (436 GC/Scion SQ). I was just wondering about everyone's experience with these instruments and what would be the better preference? Which instrument would stand out better?